NetForum uses cookies to ensure that we give you the best experience on our website. If you continue to use the site, we'll assume that you are happy to receive these cookies on the NetForum website. Read about our cookies.
NetForum Community
Learn. Share. Optimize.
Log in | Sign up now | Submit content | Contact

Attention valued NetForum members:

Due to evolving technology needs and global privacy regulations, we have made the hard decision to suspend the NetForum User Community platform on Friday, November 29, 2019.

After this date, the current NetForum can no longer be reached. Please click here for more information about this decision, what happens next and how to stay in touch with us about the future of the community.

Go to similar content

Black Blood imaging of HIV patient with brain vasculitis

Case Study
Sadeghi, Niloufar, M.D., Ph.D. Erasme Hospital • Belgium

Patient history

A 56-year-old patient presented in the Emergency Room at Erasme Hospital in Brussels, Belgium, with recurrent left leg weakness that had been occurring over a period of 24 hours. The patient was known to have been HIV infected for four years, but was not treated for this infection. The patient had multiple cardiovascular risk factors such as obesity, glucose intolerance, arterial hypertension and hyper cholesterolemia. The neurological examination showed left leg hemiparesis.

MR examination

After a conventional routine MR imaging examination, the suspicion of vasculitis arose, therefore we performed an MRI including Black Blood imaging in a separate session. The dedicated ExamCard includes diffusion, FLAIR, MR angiography using TOF, and 3D T1 MRA with bolus injection. This ExamCard also includes Black Blood imaging before and after contrast.

 

This examination was performed on our Ingenia 3.0T.

Black Blood scan time 4:39 min, acquired voxel size 0.75 x 0.75 x 1.0 mm, 21 slices.

FLAIR and DWI images

 <br>

On FLAIR images we can see some nonspecific high signal abnormalities in frontal white matter bilaterally. On DWI we can see acute ischemic lesions which appear with high signal intensity.

Pre-contrast Black Blood images

Post-contrast Black Blood imaging

 <br> <br>


Arrows show vessel wall enhancement which appears concentric and homogeneous in different cerebral territories.

One month later - post-contrast Black Blood imaging

 <br> <br>


After one month of treatment, post-contrast Black Blood images at the exact same levels as in the figure above show disappearance of the vessel wall enhancements which were seen on the previous examination.

Discussion of findings

On the routine MR sequences that we did, we could see acute ischemic lesions. We see them very well on the diffusion images, where acute ischemic lesions usually appear with high signal intensity and restricted diffusion. However, the etiology of these lesions cannot be derived from these images.

 

An area of restricted diffusion was seen in the anterior cerebral artery territory and we concluded it was an ischemic lesion. On MR angiography we can just see if there is stenosis or vessel occlusion, but it does not provide us information on the etiology of this kind of lesion.

  

So, we decided to perform Black Blood imaging. The presence and the pattern of vessel wall enhancement on Black Blood imaging, can help us to determine the etiology of the lesion.

 

Many studies have shown that Black Blood imaging can help differentiate vasculitis from other causes of vasculopathy, such as atherosclerosis, with a high specificity [1-3]. In an atherosclerotic lesion, vessel wall thickening and enhancement are usually eccentric, while in vasculitis the wall thickening and enhancement are usually concentric, homogenous, and in a long portion of the vessel.

 

Furthermore, this imaging can also be used for the follow-up of patients whenever their treatment is installed in order to determine the efficacy of a particular treatment..

 

In this case the Black Blood imaging helped us to suggest the diagnosis of HIV-related brain vasculitis.

Impact of Black Blood imaging for this patient

With the multiple cardiovascular risk factors this patient had, such as glucose intolerance, arterial hypertension and hypocholesteremia, his lesions could be atherosclerotic lesions or vasculitis, conditions which require different treatment. Especially in this patient with HIV infection causing the vasculitis, treatment of the two conditions is different.

 

The results of MRI with Black Blood imaging, helped to choose the preferred treatment for this patient, which was based on antiviral medication rather than an antiaggregant or anticoagulation treatment which is usually given to patients with risk of ischemia based on atherosclerotic lesions.

 

One month after beginning the antiviral treatment, the same MRI examination was repeated and again 8 months after the beginning of treatment. On follow-up images, we see the enhancements have almost disappeared.

  

So in case of this patient, the MRI exam with Black Blood imaging helped us to give the patient the appropriate treatment and also allowed us to noninvasively confirm the treatment response.

The importance of Black Blood imaging

Black Blood imaging can help us to noninvasively visualize vessel wall thickening and enhancement patterns that occur in vasculitis, and help us distinguish it from atherosclerotic lesions. Imaging techniques such as time-of-flight (TOF) MR angiography are not very sensitive or specific for this kind of lesions. Other possible diagnostic methods are intra-arterial angiography or brain biopsies which are both invasive.

Recommendations for using Black Blood imaging

We do not perform this examination with Black Blood imaging on all patients with ischemic lesions in the brain, because in most patients the lesion origin is embolic or atherosclerotic. We typically use it in young patients (less than 60 years old) or those patients without cardiovascular risk factors. We find it important to use Black Blood imaging in such cases, because treatment is different for a patient with vasculitis.

References

1. Swartz RH, Bhuta SS, Farb RI, Agid R, Willinsky RA, Terbrugge KG, et al. Intracranial arterial wall imaging using high-resolution 3-tesla contrast-enhanced MRI. Neurology. 2009 Feb 17;72(7):627–34.

 

2. Obusez EC, Hui F, Hajj-Ali RA, Cerejo R, Calabrese LH, Hammad T, et al. Highresolution MRI vessel wall imaging: spatial and temporal patterns of reversible cerebral vasoconstriction syndrome and central nervous system vasculitis. AJNR Am J Neuroradiol. 2014 Aug;35(8):1527–32.

 

3. Mossa-Basha M, Hwang WD, De Havenon A, Hippe D, Balu N, Becker KJ, et al. Multicontrast high-resolution vessel wall magnetic resonance imaging and its value in differentiating intracranial vasculopathic processes. Stroke J Cereb Circ. 2015 Jun;46(6):1567–73.

 

4. Cheron J, Wyndham-Thomas C, Sadeghi N, Naeije G. Response of Human Immunodeficiency Virus-Associated Cerebral Angiitis to the Combined Antiretroviral Therapy. Front. Neurol., 13 March 2017, doi.org/10.3389/fneur.2017.00095

 

 


Black Blood imaging

Philips Black Blood imaging is 3D brain imaging with reduced intraluminal blood signal1 over the complete imaging volume in the brain.

 

It helps you to better differentiate intraluminal blood signal from other signal, which can enhance diagnostic confidence.

 

The Black Blood sequence allows

• fast2, isotropic 3D imaging

• higher spatial resolution3

• reformatting in any plane without loss of resolution

 

 

1 Compared to our 3D T1W scan without MSDE prepulse

2 Compared to our 2D double inversion recovery methods with same full brain coverage

3 Compared to our 2D double inversion recovery methods with same brain coverage and scan time


 

 

 

 

Results from case studies are not predictive of results in other cases. Results in other cases may vary.



This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
952
Added:
Jun 6, 2017

Rate this:
Log in to vote
 

Case Study
Ingenia 3.0T
Black blood, Black Blood imaging, Brain, brain vasculitis, HIV, Neuro, vasculitis, vessels
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor