SENSE NeuroVascular coil

Application Tip
Rijckaert, Yvonne Philips Healthcare Philips Global
van den Heuvel, Martina Philips Healthcare Philips Global

Type of coil

The SENSE NeuroVascular coil (SENSE NV coil) is a rigid volume coil consisting of a coil base and a head portion.

 

Technical characteristics:

-> Synergy detection      -> Receive-only coil

-> SENSE compatible      -> CLEAR compatible

SENSE NV coil Coil baseSENSE NV coil Coil base with head portion attachedSENSE NV coil Dedicated headset
SENSE NV coil
SENSE NV coil
SENSE NV coil
Coil base
Coil base with head portion attached
Dedicated headset

Dedicated headset for SENSE NV coil

As the SENSE NV coil is not compatible with the head set of the Q-Head coil, there is a dedicated headset for use with the SENSE NV coil. Its hearing protection is lower than with the Q-Head coil headset. The use of earplugs is mandatory on 3.0T and strongly advised on 1.5T in combination with this dedicated headset.

An alternative to the headset are the gray wig cushions which can be placed between the patient's head and the border of the coil.

Coil positioning

  • Place the coil base on the tabletop.
  • Place the patient on the coil base.  
  • Attach the head portion to the coil base.

 

In order to facilitate isocenter positioning, the SENSE NeuroVascular coil has two markers: one in the middle of the coil in FH direction and one in the middle of the head portion.

SENSE NV coil Coil base and head portion on tabletopSENSE NV coil Attaching of head portion with patient positioned on coil base.SENSE NV coil Head portion attached to coil base
SENSE NV coil
SENSE NV coil
SENSE NV coil
Coil base and head portion on tabletop
Attaching of head portion with patient positioned on coil base.
Head portion attached to coil base

Coil elements

The SENSE NV coil consists of 18 coil elements:

  • 8 head elements,
  • 4 neck elements,
  • 6 chest elements.
SENSE NV coil elements From left to right:
All 18 coil elements \ 8 head elements \ 4 neck elements \ 6 chest elements
SENSE NV coil elements
From left to right: All 18 coil elements \ 8 head elements \ 4 neck elements \ 6 chest elements

From left to right:

All 18 coil elements      |   8 head elements       |    4 neck elements      |    6 chest elements

 

Coil element combinations

Combining the coil elements in two different ways results in the two SENSE NeuroVascular coils:

  • the 8-channel SENSE NV coil
  • the 16-channel SENSE NV coil



8-CHANNEL SENSE NV COIL

The 8-channel SENSE NV coil has two operating modes: the Head mode and the NV (NeuroVascular) mode. The figures below illustrate these modes.

 

Head mode:

The signal from each individual head coil element is directly routed to the Achieva Freewave.

This mode is preferred for head and brain imaging including intracranial MRA.

It provides excellent detail for small anatomical structures in the head.

 

NV mode:

The signals from the individual elements are combined:

  • 8 head elements to    3 channels,
  • 4 neck elements to    2 channels,
  • 6 chest elements to   3 channels

                                     --------------

      -> resulting in an  8-channel NV mode

 

This mode is preferred for cervical spine, neck imaging including extracranial MRA.
The maximal recommended field of view (FOV) is 44 cm.

 

Note that the combiner is an intrinsic part of the coil.

 

ApTip SENSE NV 8ch head&NV mode
ApTip SENSE NV 8ch head&NV mode

 

Imaging parameters with respect to coil selection

 

  • Coil selection: SENSE NV-8
  • Coil mode: Head, NV (NeuroVascular)

 


16-CHANNEL SENSE NV COIL

The signal from the individual elements is directly routed to Achieva Freewave except for the 8 head elements which are combined to 6 channels:

 

  • 8 head elements to    6 channels,
  • 4 neck elements to    4 channels,
  • 6 chest elements to   6 channels

                                     --------------

      -> resulting in a   16-channel NV coil

 

Note that the combiner is an intrinsic part of the coil.

ApTip SENSE NV coil 16 channels
ApTip SENSE NV coil 16 channels

 

Imaging parameters with respect to coil selection

 

  • Coil selection:        SENSE NV-16
  • Channel selection:  H, N, AC, PC, HN, NAC, NPC, NACPC, HNACPC

                                   where H    stands for Head,

                                             N    stands for Neck,

                                             AC  stands for Anterior Chest,

                                             PC  stands for Posterior Chest.




This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
2
Views:
9764
Added:
Apr 15, 2005

Rate this:
Log in to vote
 

Application Tip
Achieva 1.5T, Achieva 3.0T, Intera 1.5T, Intera 3.0T
Release 1, Release 10, Release 11, Release 9
Explorer / Nova Dual, Master / Nova, Master / Quasar Dual, Nova, Nova Dual, Omni / Stellar, Power / Pulsar, Pulsar, Quasar Dual
Brain, Brain stem, Cervical spine, Extracranial MRA, IAC, Neck, Neuro, Pituitary, Vascular
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor