Ingenia 1.5T knee with metal implants with and without MARS protocol

Case Study
Mr. Giordano, Kris, R.T. Barnabas Health ACC • USA

Ingenia 1.5T knee with and without using MARS protocol

In this case a knee with a metal implant is scanned. The series used for scanning are a comparison between scanning with and without the use of application tips for metal artifact reduction.

MR examination

The patient is positioned supine, feet first.
The dS Knee 16ch coil is used.

 

ExamCard used:

Images

T1W TSE COR Voxel Size 0.42 x 0.6 x 3.5 mm<br>
FOV 150 x 150 x 100 mm<br>
CLEAR, no SENSE <br>
Scan time 2.30 min.T1W TSE COR MARS Voxel size 0.42 x 0.63 x 2.5 mm<br>
FOV 150 x 150 x 99 mm<br>
Classic, no SENSE <br>
Scan time 3.21 min.
T1W TSE COR
T1W TSE COR MARS
Voxel Size 0.42 x 0.6 x 3.5 mm
FOV 150 x 150 x 100 mm
CLEAR, no SENSE
Scan time 2.30 min.
Voxel size 0.42 x 0.63 x 2.5 mm
FOV 150 x 150 x 99 mm
Classic, no SENSE
Scan time 3.21 min.
T2W TSE SAG Voxel size 0.45 x 0.58 x 3 mm<br>
FOV 150 x 150 x 89 mm<br>
CLEAR, no SENSE <br>
Scan time 3.48 min.T2W TSE SAG MARS Voxel size 0.45 x 0.56 x 2.5 mm<br>
FOV 150 x 150 x 90 mm<br>
Classic, no SENSE <br>
Scan time 4.12 min.
T2W TSE SAG
T2W TSE SAG MARS
Voxel size 0.45 x 0.58 x 3 mm
FOV 150 x 150 x 89 mm
CLEAR, no SENSE
Scan time 3.48 min.
Voxel size 0.45 x 0.56 x 2.5 mm
FOV 150 x 150 x 90 mm
Classic, no SENSE
Scan time 4.12 min.
T2W SPAIR AX Voxel size 0.72 x 0.94 x 3.5 mm<br>
FOV 150 x 150 x 131 mm<br>
CLEAR, no SENSE <br>
Scan time 4.21 min.T2W SPAIR AX MARS Voxel size 0.72 x 0.95 x 2.5 mm<br>
FOV 150 x 150 x 132 mm<br>
Classic, no SENSE <br>
Scan time 5.33 min.
T2W SPAIR AX
T2W SPAIR AX MARS
Voxel size 0.72 x 0.94 x 3.5 mm
FOV 150 x 150 x 131 mm
CLEAR, no SENSE
Scan time 4.21 min.
Voxel size 0.72 x 0.95 x 2.5 mm
FOV 150 x 150 x 132 mm
Classic, no SENSE
Scan time 5.33 min.
STIR aTSE COR MARS Voxel size 0.59 x 0.88 x 2.5 mm<br>
FOV 160 x 160 x 108 mm<br>
Classic, no SENSE <br>
Scan time 4.55 min.STIR aTSE AX MARS Voxel size 0.59 x 0.84 x 2.5 mm<br>
FOV 150 x 150 x 132 mm<br>
Classic, no SENSE <br>
Scan time 6.15 min.
STIR aTSE COR MARS
STIR aTSE AX MARS
Voxel size 0.59 x 0.88 x 2.5 mm
FOV 160 x 160 x 108 mm
Classic, no SENSE
Scan time 4.55 min.
Voxel size 0.59 x 0.84 x 2.5 mm
FOV 150 x 150 x 132 mm
Classic, no SENSE
Scan time 6.15 min.



Tips and hints to reduce metal artifacts in sequences: 
  • Use TSE instead of FFE (also for the survey)
  • Use Uniformity Classic
  • Use short echo spacing and high TSE factor
  • Use a small WFS of 0.5 to 0.7
  • Use thin slices
  • Maintain good SNR (give up some in-plane resolution)
  • Use STIR, no spectral fat saturation


If you have to use a gradient echo the principles as outlined in the previous tips still apply.



Disclaimer:
Metal implants are a contraindication for MRI, unless the MR compatibility for the implant is stated by the implant manufacturer. MR healthcare professionals are advised to contact the respective implant manufacturer in order to obtain the latest safety information to ensure patient safety relative to the use of an MR procedure.


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
1
Views:
1955
Added:
Jun 18, 2013

Rate this:
Log in to vote
 

Case Study
Ingenia 1.5T
Release 4.1
dS Knee coil, dStream, implant, Knee, MARS, metal artifact reduction, Musculoskeletal, prosthesis
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor