Metal artifact reduction for MRI of metal prostheses and implants

Application Tip
van Meel, Marius Philips Healthcare • Netherlands

Patients suffering from pain after receiving hip or knee implants may have soft tissue disease. MRI is ideal for soft tissue sensitivity, but the metal implant causes magnetic field disturbances leading to an artifact which complicates image interpretation. Metal artifact reduction sequences (MARS) are less sensitive to metal artifacts. These tips show how to adapt your scans to reduce the metal artifacts. ExamCards will be available on NetForum shortly.

Tip 1: Use TSE instead of FFE

TSE will show less pronounced metal artifacts than gradient echo sequences (FFE and TFE), because of the refocusing pulses used in TSE.

 TFE: strong metal  artifact TSE: less metal atrifact
TFE: strong metal artifact
TSE: less metal atrifact


So, use preferably TSE sequences instead of FFE. Also use a TSE survey – an example can be found in the IAC preset protocols:

Tip 2: SENSE or CLEAR

Switch off SENSE and CLEAR, no in the Geometry tab. Without SENSE or CLEAR, the reference

scan can also be removed from the ExamCard to save scan time.

 improve picture!
improve picture!

Tip 3: Use short echo spacing and high TSE factor

Nowadays, asymmetric TSE allows more direct control of ES, and the TSE factor can be

increased to speed up the scan. Don’t make the TSE shot larger than about 4 x TE to avoid

blurring. In older releases with linear profile order, a short echo spacing (ES) could be only

obtained by setting a high TSE factor.

Tip 4: Use a small WFS of 0.5 to 0.7

Use a small WFS of 0.5 to 0.7; this corresponds to a high bandwidth.

For more explanation, see application tip in FieldStrength issue 36, Dec. 2008.

These low WFS values require strong gradients, so use gradient mode maximum.

If you still can’t reach the low WFS, then you may set it to minimum. It is not recommended to do this from the beginning, as WFS may become lower than needed, which leads to low SNR.

 WFS 2, bandwidth 110 Hz/pixel WFS 0.5, bandwidth 440 Hz/pixel
WFS 2, bandwidth 110 Hz/pixel
WFS 0.5, bandwidth 440 Hz/pixel

Tip 5: Use thin slices

Selecting thin slices will need a strong slice selection gradient, which is similar to setting a high bandwidth for the readout, and thus will decrease through-plane artifacts, as seen in this example.

 Voxels 0.8 x 1.0 x 6 mm Voxels 0.8 x 1.0 x 3 mm
Voxels 0.8 x 1.0 x 6 mm
Voxels 0.8 x 1.0 x 3 mm

TIP 6: Maintaining good SNR

Increase NSA to compensate when measures of the previous tips cost some SNR.
The typical scan time of a MARS protocol is about 5 minutes.
If you want to further increase SNR, give up some in-plane resolution first.

 Standard PDW TSE MARS - everything together
Standard PDW TSE
MARS - everything together


The right image shows the result of applying all tips: PDW TSE with asymmetric profile order, ES 6 ms, TSE factor 20, WFS 0.5, voxels 0.8 x 1.0 x 2.0 mm, no SENSE, no CLEAR, NSA 3, scan time 4:54 min. This is done on Achieva 1.5T with 32-channel SENSE Cardiac coil.

TIP 7: Use STIR, no spectral fat saturation

Use STIR for fat suppression. This is implemented in the MARS ExamCards on NetForum, fitting the scan in one package for 1.5T or two packages for 3.0T. Do not use spectral fat saturation such

as SPIR or SPAIR, because these techniques perform best in a homogeneous magnetic field.

 SPAIR SPAIR
SPAIR
SPAIR
 STIR STIR
STIR
STIR

TIP 8: Gradient echo images

Sometimes, gradient echo (FFE or TFE) may be needed and cannot be replaced by TSE.

Note that in such cases principles as outlined in the previous tips still apply.


View ExamCard:




Disclaimer

Metal implants are a contraindication for MRI, unless the MR compatibility for the implant is stated by the implant manufacturer. MR healthcare professionals are advised to contact the respective implant manufacturer in order to obtain the latest safety information to ensure patient safety relative to the use of an MR procedure.

 

 

View additional information:

www.revisemri.com/blog/2011/metal-artefact-reduction/

 

 





This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
9
Views:
25240
Added:
Oct 25, 2011

Rate this:
Log in to vote
 

Application Tip
Achieva 1.5T, Achieva 3.0T
Hip, Knee, MARS, metal artifact reduction, Musculoskeletal, prosthesis
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor