SENSE factor and direction in shoulder and knee imaging

Application Tip
Teitsma, Loretta Philips Healthcare Philips Global
van den Heuvel, Martina Philips Healthcare Philips Global

Knee imaging

  • Used coil: SENSE Flex-M or SENSE Flex-S coil
  • It is best to position the two coil elements anterior and posterior
  • The preferred SENSE direction (for optimal SNR) is orthogonal to the coil elements (AP)

 

Sagittal image orientation:

SENSE direction: AP optimal, FH possible (strange noise patterns)

SENSE factor: no more than 1.5 (noise breakthrough)

 

Coronal image orientation:

SENSE direction: LR possible but not optimal (strange noise patterns )

SENSE factor: no more than 1.5 (noise breakthrough)

 

Axial image orientation:

SENSE direction: AP optimal

SENSE factor: 2.0 possible with 2 NSA

Fold-over suppression to avoid intrinsic fold-over

 

Scan times: 2-3 minutes.

Shoulder imaging

  • Used coil: SENSE Flex-M or SENSE Flex-L coil
  • Position the coil elements as described in the Application Tip "SENSE Flex-M coil: positioning for a shoulder examination".

 

Coronal image orientation:

SENSE direction: FH as well as LR are possible, but not optimal.

 

Sagittal image orientation:

SENSE direction: AP is optimal.

SENSE factor: 2 is possible.

Note: be careful in angulating to avoid intrinsic fold-over.

 

Axial image orientation:

SENSE direction: AP is optimal.

SENSE factor: 2 is possible.

Use 2 NSA and fold-over suppression.

NOTE: be careful in angulating to avoid intrinsic fold-over.

 

Scan times: 2-3 minutes.

Links to related documents

This Application Tip "SENSE Flex coils: applications and positioning" provides

 

  • Information on the coils, on their applications and on general coil positioning
  • More links to other related documents, all dealing with the SENSE Flex coils


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
1
Views:
5464
Added:
Nov 25, 2004

Rate this:
Log in to vote
 

Application Tip
Achieva 1.5T, Achieva 3.0T, Intera 1.0T, Intera 1.5T, Intera 3.0T
Release 1, Release 10, Release 11, Release 2, Release 2.5, Release 2.6, Release 9
Explorer / Nova Dual, gradients, Master / Nova, Master / Quasar Dual, Nova, Nova Dual, Omni / Stellar, Omni / Stellar, Power / Pulsar, Pulsar, Quasar, Quasar Dual
Knee, Musculoskeletal, Shoulder
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor