0.6T Axilla - CUH, Herlev

ExamCard
Larsen, Lone Herlev, Copenhagen University Hospital Denmark
Wirtanen, Pirjo Philips Healthcare Philips Global

0.6T CIA Axilla - CUH, Herlev

Download
Login is required to download ExamCards.

ExamCard purpose

This ExamCard is for imaging the axillary area using CIA to combine the Flex M+ and Flex L+ coils. The resulting coil combination has a large coverage that enables using a large field of view or large number of slices.

 

SENSE is used to shorten the imaging time.

 

T2w TSE Tra has a large coverage in the slice direction (48 slices).

PDw STIR has a large field of view (380 mm). NSA 2 is used to suppress the effects of axillary flow.

 

Total imaging time is about 20 minutes.

ExamCard overview

CIA Axilla
Scan 13 Plane scoutview
Scan 2Coronal scoutview
Scan 3T2w TSE Tra
Scan 4STIR TSE Cor
Scan 5T1w TSE Tra
Scan 6IV contrast
Scan 7T1w FS-TSE Cor
Scan 8T1w FS-SE Tra

ExamCard images

T2w TSE transverse FOV 260<br>Matrix 300<br>Slice thickness 5.0 mm<br>SENSE factor 2.0<br>
Scan time 3:44 min.T1w TSE transverse FOV 260<br>Matrix 614<br>Slice thickness 5.0 mm<br>SENSE factor 2.0<br>
Scan time 3:10 min.T1w FS-SE transversee FOV 300<br>Matrix 308<br>Slice thickness 6.0 mm<br>
Scan time 3:48 min.
T2w TSE transverse
T1w TSE transverse
T1w FS-SE transversee
FOV 260
Matrix 300
Slice thickness 5.0 mm
SENSE factor 2.0
Scan time 3:44 min.
FOV 260
Matrix 614
Slice thickness 5.0 mm
SENSE factor 2.0
Scan time 3:10 min.
FOV 300
Matrix 308
Slice thickness 6.0 mm
Scan time 3:48 min.
PDw STIR coronal FOV 380<br>Matrix 288<br>Slice thickness 6.0 mm<br>SENSE factor 2.0<br>
Scan time 3:10 min.Slice positioning for transverse plane
PDw STIR coronal
Slice positioning for transverse plane
FOV 380
Matrix 288
Slice thickness 6.0 mm
SENSE factor 2.0
Scan time 3:10 min.

Patient and coil positioning

Position the coils for CIA perpendicularly to each other to apply SENSE. If smaller coverage is needed, coils can be positioned closer to each other.

 

CIA Coil positioning Positioning the coils perpendicular to each other enables use of SENSE in all planes.
CIA Coil positioning
Positioning the coils perpendicular to each other enables use of SENSE in all planes.


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
1322
Added:
Dec 13, 2006

Rate this:
Log in to vote
 

ExamCard
Panorama 0.6T
Release 6
Standard
Musculoskeletal, Shoulder
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor