0.6T Brain with Neck L coil for claustrophobic patients

ExamCard
Wirtanen, Pirjo Philips Healthcare Philips Global

Download
Login is required to download ExamCards.

ExamCard purpose

This ExamCard can be used for a brain examination. The Neck L coil is used inversely, so that there are no mechanical parts above the patients┬┤ face, creating a more open feel.

 

This method is in use in the University Hospital Herlev in Copenhagen, Denmark.

 

This ExamCard consists of T2w, T1w, FLAIR, and Inflow scans. Alternative scans can be taken from Philips Nice Brain or Excellent Brain protocols, if longer imaging times can be accepted. 

If very short scans are needed, take additional scans from Philips Fast Brain and Speedy Brain protocols.

 

If diffusion imaging is needed, DWISE and TDWISE are recommended. Those can be found in the Philips Speedy Brain and Philips Diffusion protocols.

ExamCard overview

Brain for Neck L coil
Scan 1Scoutview
Scan 2T2w TSE Tra
Scan 3FLAIR Tra
Scan 4T2w TSE Sag
Scan 5T1w FFE3D Tra
Scan 6T1w SE Tra
Scan 7Inflow
Scan 8FAST T2w TSE Tra
Scan 9FAST FLAIR Sag
Scan 10FAST T2w TSE Sag
Scan 11FAST T1w FFE3D Tra
Scan 12FAST T1w SE Tra
Scan 13FAST Inflow

ExamCard images

T2w TSE transverse FOV 240, Slice thickness 4.0 mm, Matrix 432,
Scan time 3:07 min.T2w TSE sagittal FOV 250, Slice thickness 4.0 mm, Matrix 346,
Scan time 4:53 min.FLAIR transverse FOV 240, Slice thickness 4.0 mm, Matrix 302,
Scan time 3:41 min.
T2w TSE transverse
T2w TSE sagittal
FLAIR transverse
FOV 240, Slice thickness 4.0 mm, Matrix 432, Scan time 3:07 min.
FOV 250, Slice thickness 4.0 mm, Matrix 346, Scan time 4:53 min.
FOV 240, Slice thickness 4.0 mm, Matrix 302, Scan time 3:41 min.
T1w FFE3D transverse FOV 240, Slice thickness 4.0 mm, Matrix 420,
Scan time 4:19 min.T1w FFE3D sagittal FOV 240, Slice thickness 4.0 mm, Matrix 420,
Scan time 4:19 min.FLAIR coronal FOV 240, Slice thickness 4.0 mm, Matrix 302,
Scan time 3:41 min.
T1w FFE3D transverse
T1w FFE3D sagittal
FLAIR coronal
FOV 240, Slice thickness 4.0 mm, Matrix 420, Scan time 4:19 min.
FOV 240, Slice thickness 4.0 mm, Matrix 420, Scan time 4:19 min.
FOV 240, Slice thickness 4.0 mm, Matrix 302, Scan time 3:41 min.
T1w SE transverse FOV 240, Slice thickness 4.0 mm, Matrix 334,
Scan time 3:50 minT1w SE coronal FOV 240, Slice thickness 4.0 mm, Matrix 334,
Scan time 3:50 minInflow MIP of intra cranial arteries FOV 200,  Slice thickness 0.6 mm, Matrix 630,
Scan time 10:45 min
T1w SE transverse
T1w SE coronal
Inflow MIP of intra cranial arteries
FOV 240, Slice thickness 4.0 mm, Matrix 334, Scan time 3:50 min
FOV 240, Slice thickness 4.0 mm, Matrix 334, Scan time 3:50 min
FOV 200, Slice thickness 0.6 mm, Matrix 630, Scan time 10:45 min
FAST T2w TSE transverse FOV 250, Slice thickness 5.0 mm, Matrix 308,
Scan time 2:10 min.FAST T2w TSE sagittal FOV 250, Slice thickness 5.0 mm, Matrix 310,
Scan time 3:18 min.FAST FLAIR transverse FOV 250, Slice thickness 5.0 mm, Matrix 312,
Scan time 2:30 min.
FAST T2w TSE transverse
FAST T2w TSE sagittal
FAST FLAIR transverse
FOV 250, Slice thickness 5.0 mm, Matrix 308, Scan time 2:10 min.
FOV 250, Slice thickness 5.0 mm, Matrix 310, Scan time 3:18 min.
FOV 250, Slice thickness 5.0 mm, Matrix 312, Scan time 2:30 min.
FAST T1w FFE3D transverse FOV 250, Slice thickness 5.0 mm, Matrix 384,
Scan time 1:45 min.FAST T1w SE transverse FOV 250, Slice thickness 5.0 mm, Matrix 334,
Scan time 2:09 min.FAST Inflow MIP of intra cranial arteries FOV 200, Slice thickness 0.6 mm, Matrix 576,
Scan time 5:36 min.
FAST T1w FFE3D transverse
FAST T1w SE transverse
FAST Inflow MIP of intra cranial arteries
FOV 250, Slice thickness 5.0 mm, Matrix 384, Scan time 1:45 min.
FOV 250, Slice thickness 5.0 mm, Matrix 334, Scan time 2:09 min.
FOV 200, Slice thickness 0.6 mm, Matrix 576, Scan time 5:36 min.

Positioning

Use two thin pads under the patient's upper torso and the standard Neck L coil support under the coil. Coil position is inversed to the coil support, the outlet of the coil cable is towards the patient. Shield the coil cable with the pads. Use enough accessories to ensure patient comfort.

 

Position the patient deep into the Neck L coil and ensure that the whole brain is covered by the coil loop. Center the laser to the eyebrows.

 Coil positioning on the patient support. Coil positioning on the patient support. Patient positioning.
Coil positioning on the patient support.
Coil positioning on the patient support.
Patient positioning.


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
1210
Added:
Sep 7, 2006

Rate this:
Log in to vote
 

ExamCard
Panorama 0.6T
Release 6
Standard
Brain, Neuro
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor