0.6T CIA Panoramic L-spine and SI joint - CUH, Herlev

ExamCard
Larsen, Lone Herlev, Copenhagen University Hospital Denmark
Wirtanen, Pirjo Philips Healthcare Philips Global

0.6T CIA Panoramic lumbar spine and SI joint - CUH, Herlev

Download
Login is required to download ExamCards.

ExamCard purpose

This ExamCard uses CIA (Combined Image Aquisition) Panoramic imaging for the lumbar spine and the SI joint with Flex L+ and Flex M+ coils. The Panoramic method was introduced in R6.1 and CIA features were introduced in R6.2. This ExamCard requires both Panoramic and CIA licenses.

 

This ExamCard is used to image the SI joint, primarily for suspected ankylosing spondylitis. Increased signal on STIR images and after i.v. gadolinium increased signal in post-contrast fat sat T1w images suggest an active spondylitis.

 

SENSE is used in the sagittal plane to shorten the acquisition times.

 

Examination time is about 22 minutes in addition to patient preparation and positioning times.

Patient positioning

CIA (Combined Image Aquisition) Panoramic imaging combines the two Flex coils M+ and L+ (using both coils at the same time is CIA Imaging).

 

  • The patient is firmly positioned sideways using the vacuum mattress, looking outside (Panoramic imaging). 
  • Two special Flex M+ and L+ coils are positioned in the same-plane way behind the back of the patient.
  • Position the patient on the vacuum cushion sideways.
  • If needed, place the pad between the knees for the patient comfort.
  • If needed, use thinner mattress under the patient.
  • Place the Flex M+ and Flex L+ coil on the same plane behind the patient´s back.
  • Support the coils with the coil holders, pads or sandbags and pump the air out of the vacuum cushion.
  • Use the "low" tabletop height setting.
 Coil selection and patient orientation
Coil selection and patient orientation

ExamCard overview

Note that the plane names refer to the true scan orientation with respect to the patient, not with respect to the magnet.
Lumbar and SI
Scan 13 plane Scout
Scan 2sag T2w STIR LS
Scan 3sag T1w TSE LS
Scan 4cor T1w TSE SI
Scan 5cor T2w STIR SI
Scan 6cor T1w FS-TSE SI C+
Scan 7tra T1w FS-TSE SI C+

ExamCard images

Sagittal T2w TSE FOV 360, Matrix 302, Slice thickness 4.0 mm, SENSE Factor 1.5, Scan time 2:48 minSagittal T1w TSE FOV 400, Matrix 352, Slice thickness 4,0 mm, SENSE factor 1.5, Scan time 3:24 min
Sagittal T2w TSE
Sagittal T1w TSE
FOV 360, Matrix 302, Slice thickness 4.0 mm, SENSE Factor 1.5, Scan time 2:48 min
FOV 400, Matrix 352, Slice thickness 4,0 mm, SENSE factor 1.5, Scan time 3:24 min
Transverse T2w STIR IR-TSE FOV 280, Matrix 334, Slice thickness 4.5 mm, SENSE  factor 1.5, Scan time 1:48 minCoronal T1w TSE FOV 280, Matrix 334, Slice thickness 4.5 mm, SENSE factor 2.0, Scan time 1:48 minTransverse T1w FS-TSE FOV 280, Matrix 334, Slice thickness 4.5 mm, SENSE factor 2.0, Scan time 1:50 min
Transverse T2w STIR IR-TSE
Coronal T1w TSE
Transverse T1w FS-TSE
FOV 280, Matrix 334, Slice thickness 4.5 mm, SENSE factor 1.5, Scan time 1:48 min
FOV 280, Matrix 334, Slice thickness 4.5 mm, SENSE factor 2.0, Scan time 1:48 min
FOV 280, Matrix 334, Slice thickness 4.5 mm, SENSE factor 2.0, Scan time 1:50 min


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
1362
Added:
Sep 18, 2006

Rate this:
Log in to vote
 

ExamCard
Panorama 0.6T
Release 6
Lumbar spine, Neuro
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor