3.0T Prostate - Symbion Research Clinical Imaging Center, Australia

ExamCard
Shnier, Ron, M.D. Symbion Research Clinical Imaging Center Australia

Download
Login is required to download ExamCards.

ExamCard purpose

This release 2.5.3 Achieva 3.0T ExamCard has been created by the Symbion Research Clinical Center (Prince of Wales Research Imaging Center) in Randwick, Sydney, Australia. It provides protocols for imaging the prostate gland.

ExamCard overview

 

This ExamCard combines the SENSE Torso coil and Endo coil in a dual-coil setup.

 

High resolution T2-weighted TSE series are included in the axial, coronal and sagittal planes to assess the areas of disease.

 

The T1-weighted images are included to assess of the lymph nodes for disease.

 

Axial diffusion weighted images are acquired to show areas of restricted diffusion.

 

3D CSI is performed to confirm areas of prostate involved in the disease.

ExamCard images

T2-weighted TSE coronal with CLEAR<br>
TR: 3385 ms<br>
TE: 111 ms<br>
Nr slices: 20<br>
FOV: 160<br>
Voxel size: 0.38 x 0.41 x 2.5 mm<br>
Scan time: 4:17 min.<br>T2-weighted TSE axial with CLEAR<br>
TR: 4657 ms<br>
TE: 106 ms<br>
Nr slices: 24<br>
FOV: 220<br>
Voxel size: 0.22 x 0.32 x 3 mm<br>
Scan time: 6:12 min.<br>T1-weighted TSE TR: 484 ms<br>
TE: 8.2 ms<br>
Nr slices: 32<br>
FOV: 300<br>
Voxel size: 1.17 x 1.3 x 5 mm<br>
Scan time: 4:48min.<br>
T2-weighted TSE coronal
T2-weighted TSE axial
T1-weighted TSE
with CLEAR
TR: 3385 ms
TE: 111 ms
Nr slices: 20
FOV: 160
Voxel size: 0.38 x 0.41 x 2.5 mm
Scan time: 4:17 min.
with CLEAR
TR: 4657 ms
TE: 106 ms
Nr slices: 24
FOV: 220
Voxel size: 0.22 x 0.32 x 3 mm
Scan time: 6:12 min.
TR: 484 ms
TE: 8.2 ms
Nr slices: 32
FOV: 300
Voxel size: 1.17 x 1.3 x 5 mm
Scan time: 4:48min.
DWI SPAIR <br>
SENSE factor: 2<br>
Scan time: 7:30 min.ADC map An ADC map was manually created from the DWI series.
DWI
ADC map
SPAIR
SENSE factor: 2
Scan time: 7:30 min.
An ADC map was manually created from the DWI series.
CSI left side Spectroscopy scan time: 13:20 min.<br>
The SpectroView application was used to post-process the spectroscopy series.CSI right side Spesctroscopy scan time: 13:20 min.<br>
The SpectroView application was used to post-process the spectroscopy series.
CSI left side
CSI right side
Spectroscopy scan time: 13:20 min.
The SpectroView application was used to post-process the spectroscopy series.
Spesctroscopy scan time: 13:20 min.
The SpectroView application was used to post-process the spectroscopy series.

Patient preparation

Position the patient supine, head first with the Endo coil in situ. Inflate the balloon of the coil with perflurocarbon solution, which provides no signal on the images and aids in shimming for spectrosopy.

 

 

View Clinical Case:



This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
2751
Added:
May 14, 2009

Rate this:
Log in to vote
 

ExamCard
Achieva 3.0T X-series
Release 2.5
Quasar Dual
3T, Body, endocoil, Prostate
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor