3.0T Routine ankle, foot - PAMF, USA

ExamCard
Goumas, Chris, M.D. Palo Alto Medical Foundation USA

Download
Login is required to download ExamCards.

ExamCard purpose

This release 2.1.3 ExamCard was created by the Palo Alto Medical Foundation for routine imaging of ankle or foot with the 8-channel SENSE Foot/Ankle coil using small FOV and high matrix. Each protocol has been customized to have an acceptable scan time of approximately five minutes or less.

ExamCard overview

Scan 1Survey
Scan 2AX PD FS
Scan 3AX T1
Scan 4SAG PD FS
Scan 5SAG T1
Scan 6COR PD FS
Scan 7COR T1

 

 

Survey: images are acquired in three orthogonal planes. The ankle or foot is positioned as close to the isocenter as possible.

 

Axial, sagittal and coronal PD FS: these proton density weighted fat suppressed protocols use SPAIR, balanced with a TI to give a slight differentiation between muscle and fat tissues. These protocols also make use of an asymmetric profile order to minimize TSE blurring effects while allowing a TE value to be used for increased signal. As a result of these signal improving benefits, the matrix and FOV of each of these protocols are minimized for optimal resolution. Due to the unaffected leg being elevated, foldover suppression is not selected for the coronal sequence. This results in this protocol benefitting from a signal averaging of multiple NSAs.

 

Axial, sagittal and coronal T1: these sequences also use an asymmetric profile order to minimize potential blurring and because of the higher signal obtained, compared to their fat suppressed counter parts, a higher resolution can be produced from these protocols. However, like the coronal proton density weighted sequence, the T1W coronal sequence does not use foldover suppression and can further benefit from a signal averaging of multiple NSAs.

Example images ankle study

Survey Right ankle in 3 orthogonal planes
Survey
Right ankle in 3 orthogonal planes
Axial PD FS with SPAIR and SENSE<br>
FOV 140 mm<br>
Sl Thk 3.5/0.3<br>
Acquired matrix  0.27f x 0.35p<br>
Scan time 5:37 min.Axial T1 with SENSE<br>
FOV 140 mm<br>
Sl Thk 3.5/0.3<br>
Acquired matrix 0,21f x 0.34p<br>
Scan time 4:02 min.
Axial PD FS
Axial T1
with SPAIR and SENSE
FOV 140 mm
Sl Thk 3.5/0.3
Acquired matrix 0.27f x 0.35p
Scan time 5:37 min.
with SENSE
FOV 140 mm
Sl Thk 3.5/0.3
Acquired matrix 0,21f x 0.34p
Scan time 4:02 min.
Sagittal PD FS with SPAIR and SENSE<br>
FOV 140 mm<br>
Sl Thk 3.5/0.3<br>
Acquired matrix 0.23f x 0.51p<br>
Scan time 4:54 min.Sagittal T1 with SENSE<br>
FOV 140 mm<br>
Sl Thk 3.5/0.3<br>
Acquired matrix 0.27f x 0.42p<br>
Scan time 5:02 min.
Sagittal PD FS
Sagittal T1
with SPAIR and SENSE
FOV 140 mm
Sl Thk 3.5/0.3
Acquired matrix 0.23f x 0.51p
Scan time 4:54 min.
with SENSE
FOV 140 mm
Sl Thk 3.5/0.3
Acquired matrix 0.27f x 0.42p
Scan time 5:02 min.
Coronal PD FS with SPAIR and SENSE<br>
FOV 140 mm<br>
Sl Thk 3.5/0.3<br>
Acquired matrix 0.23f x 0.51p<br>
Scan time 4:59 min.Coronal T1 with SENSE<br>
FOV 140 mm<br>
Sl Thk 3.5/0.3<br>
Acquired matrix 0.21f x 0.34p<br>
Scan time 4:03 min.
Coronal PD FS
Coronal T1
with SPAIR and SENSE
FOV 140 mm
Sl Thk 3.5/0.3
Acquired matrix 0.23f x 0.51p
Scan time 4:59 min.
with SENSE
FOV 140 mm
Sl Thk 3.5/0.3
Acquired matrix 0.21f x 0.34p
Scan time 4:03 min.

 

 

 

Example images foot study

Note: the ExamCard has been adapted for the foot study.
Survey Left foot in 3 orthogonal planes.
Survey
Left foot in 3 orthogonal planes.
Coronal PD FS With SPAIR and SENSE<br>
FOV; 140 mm<br>
Sl Thk: 3.0/0.03Sagittal PD FS With SPAIR and SENSE<br>
FOV: 140 mm<br>
Sl Thk: 3.0/0.3AX PD FS With SPAIR and SENSE<br>
FOV: 100 mm<br>
Sl Thk: 3.0/0.3
Coronal PD FS
Sagittal PD FS
AX PD FS
With SPAIR and SENSE
FOV; 140 mm
Sl Thk: 3.0/0.03
With SPAIR and SENSE
FOV: 140 mm
Sl Thk: 3.0/0.3
With SPAIR and SENSE
FOV: 100 mm
Sl Thk: 3.0/0.3
Coronal T1W With SENSE<br>
FOV; 140 mm<br>
Sl Thk: 3.0/0.03Sagittal T1W With SENSE<br>
FOV; 140 mm<br>
Sl Thk: 3.0/0.03AX T1W With SENSE<br>
FOV; 100 mm<br>
Sl Thk: 3.0/0.03
Coronal T1W
Sagittal T1W
AX T1W
With SENSE
FOV; 140 mm
Sl Thk: 3.0/0.03
With SENSE
FOV; 140 mm
Sl Thk: 3.0/0.03
With SENSE
FOV; 100 mm
Sl Thk: 3.0/0.03

Patient preparation

For both studies the patient is positioned feet first, supine, with the ankle or foot placed in the isocenter as much as possible. The unaffected leg is elevated and supported with padding for comfort along the side of the 8-channel SENSE Foot/Ankle coil. Elevating the leg enables the coronal protocols to be acquired without having to use foldover suppression.

 

 

For more information:



This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
1633
Added:
Jun 15, 2008

Rate this:
Log in to vote
 

ExamCard
Achieva 3.0T X-series
3T, Ankle, Musculoskeletal
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor