HFO Elbow with ST Extremity coil and SENSE

ExamCard
Wirtanen, Pirjo Philips Healthcare Philips Global

Download
Login is required to download ExamCards.

ExamCard purpose

This release 2.1 Panorama HFO ExamCard provides a collection of protocols for elbow imaging. The protocols are optimized for use with the ST Extremity coil using SENSE and are arranged according to image contrast.

 

SENSE is used to obtain higher resolution with scan times of approximately 5 minutes per protocol. The coil loops are positioned in a perpendicular fashion (see photos below), which enables SENSE to be used in all directions.

ExamCard overview

Elbow with the ST Extremity coil and SENSE
Scan 1Survey
Scan 2Ref Scan
Scan 3T1w TSE
Scan 4T1w SPIR
Scan 5T2w TSE
Scan 6T2w SPAIR
Scan 7PDw DRIVE
Scan 8PDw DR SPAIR
Scan 9STIR LongTE
Scan 10WATSc
Scan 11WATSf

ExamCard images

T2w TSE axial FOV 130 x 110 mm<br>
Voxel size 0.47 x 0.51 mm<br>
Slice thickness 3.0 mm<br>
SENSE factor 1.4<br>
Scan time 4:13 min.T2w SPAIR coronal FOV 130 x 110 mm<br>
Voxel size 0.55 x 0.68 mm <br>
Slice thickness 3.0 mm<br>
SENSE factor 1.6<br>
Scan time 4:54 min.WATSf sagittal FOV 130 x 110 mm<br>
Voxel size 0.47 x 0.47 mm<br>
Slice thickness 1.5 mm<br>
SENSE factor 1.5<br>
Scan time 4:38 min.
T2w TSE axial
T2w SPAIR coronal
WATSf sagittal
FOV 130 x 110 mm
Voxel size 0.47 x 0.51 mm
Slice thickness 3.0 mm
SENSE factor 1.4
Scan time 4:13 min.
FOV 130 x 110 mm
Voxel size 0.55 x 0.68 mm
Slice thickness 3.0 mm
SENSE factor 1.6
Scan time 4:54 min.
FOV 130 x 110 mm
Voxel size 0.47 x 0.47 mm
Slice thickness 1.5 mm
SENSE factor 1.5
Scan time 4:38 min.
T1w TSE coronal FOV 130 x 110 mm<br> Voxel size 0.42 x 0.47mm<br>
Slice thickness 3.0 mm<br> SENSE Factor 1.8<br>
Scan time 4:14 min.T1w SPIR axial FOV 130 x 110<br>
Voxel size 0.57 x 0.71 mm<br>
Slice thickness 3.0 mm<br>
SENSE factor 1.6<br>
Scan time 3:59 min.WATSc axial FOV 130 x 110 mm<br>
Voxel size 0.47 x 0.47mm<br>
Slice thickness 1.5 mm<br>
SENSE factor 2.0<br>
Scan time 4:26 min.
T1w TSE coronal
T1w SPIR axial
WATSc axial
FOV 130 x 110 mm
Voxel size 0.42 x 0.47mm
Slice thickness 3.0 mm
SENSE Factor 1.8
Scan time 4:14 min.
FOV 130 x 110
Voxel size 0.57 x 0.71 mm
Slice thickness 3.0 mm
SENSE factor 1.6
Scan time 3:59 min.
FOV 130 x 110 mm
Voxel size 0.47 x 0.47mm
Slice thickness 1.5 mm
SENSE factor 2.0
Scan time 4:26 min.
PDw DRIVE coronal FOV 130 x 110 mm<br>
Voxel size 0.40 x 0.48 mm <br>
Slice thickness 3.0 mm<br>
SENSE factor 1.7<br>
Scan time 3:41 min.PDw DRIVE SPAIR axial FOV 130 x 110<br>
Voxel 0.55 x 0.68 mm<br>
Slice thickness 3.0 mm<br>
SENSE factor 1.5<br>
Scan time 4:57 min.STIR LongTE sagittal FOV 130 x 110<br>
Voxel size 0.61 x 0.77 mm <br>
Slice thickness 3.0 mm<br>
SENSE factor 1.6<br>
Scan time 4:26 min.
PDw DRIVE coronal
PDw DRIVE SPAIR axial
STIR LongTE sagittal
FOV 130 x 110 mm
Voxel size 0.40 x 0.48 mm
Slice thickness 3.0 mm
SENSE factor 1.7
Scan time 3:41 min.
FOV 130 x 110
Voxel 0.55 x 0.68 mm
Slice thickness 3.0 mm
SENSE factor 1.5
Scan time 4:57 min.
FOV 130 x 110
Voxel size 0.61 x 0.77 mm
Slice thickness 3.0 mm
SENSE factor 1.6
Scan time 4:26 min.

Patient positioning

The objective is to position the coil loops very carefully perpendicular to each other. Position the first coil loop around the elbow over the joint, with the handle under the elbow. The second coil loop should then be placed close to the side of the elbow inside the first loop, with the handle on the top (see the photo below). 

 

Another option is to position the coils perpendicular to each other first and then place the elbow  in the middle as much as possible.

 

Use sand bags to stabilize the coil and for patient comfort.

 

Make sure that the coil cables do not create any loops and that there is no skin contact with any part of the coil.

Coil position Position one coil loop around the elbow and the other as close to the side of the elbow as possible.Optional coil position First place the coils in a perpendicular position and then place the elbow in the middle. Keep the coil loops at 90 degrees to each other.Secure coil cables Use sand bags to stabilize the coils and to prevent the cables from forming loops.
Coil position
Optional coil position
Secure coil cables
Position one coil loop around the elbow and the other as close to the side of the elbow as possible.
First place the coils in a perpendicular position and then place the elbow in the middle. Keep the coil loops at 90 degrees to each other.
Use sand bags to stabilize the coils and to prevent the cables from forming loops.


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
1496
Added:
Sep 26, 2008

Rate this:
Log in to vote
 

ExamCard
Panorama HFO
Release 2
Pulsar
1.0T, 1T, Elbow, Musculoskeletal
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor