HFO Shoulder with the ST Extremity coil and SENSE

ExamCard
Wirtanen, Pirjo Philips Healthcare Philips Global

Download
Login is required to download ExamCards.

ExamCard purpose

This release 2.1 Panorama HFO ExamCard has been created for shoulder imaging using the ST Extremity coil and SENSE. The advantage of using the ST Extremity coil is that SENSE can be applied in all scanning directions, enabling the scan times for each protocol to be approximately 3 minutes or less.

 

The coverage and scan times are adapted to be used with 20 slices.

 

It is recommended to use the fold-over option with the direction for coronal being R-L, for transverse A-P and for sagittal A-P

 

ExamCard overview

Shoulder with the ST Extremity coil and SENSE
Scan 1Survey
Scan 2Ref Scan
Scan 3T1w TSe
Scan 4T1w SPIR
Scan 5T2w TSE
Scan 6T2w SPAIR
Scan 7T2*w FFE InPhase
Scan 8T2*w FFE OutPhase
Scan 9PDw DRIVE
Scan 10PDw DR SPAIR
Scan 11STIR LongTE
Scan 12WATSc
Scan 13WATSf

ExamCard Images

T1w TSE transverse FOV 150 mm<br>
Voxel size 0.75 x 0.86 mm,<br>
Slice thickness 3.5 mm<br>
SENSE factor 1.5<br>
Scan time 2:47 min.T1w SPIR coronal FOV 150 mm<br>
Voxel size 0.75 x 0.96 mm<br>
Slice thickness 3.5 mm<br>
SENSE factor 1.5<br>
Scan time 2:17 min.T2w FFE OutPhase coronal FOV 150 mm<br>
Voxel size 0.65 x 0.77 mm<br>
Slice thickness 1.5 mm<br>
SENSE factor 1.6 <br>
Scan time 2:07 min.
T1w TSE transverse
T1w SPIR coronal
T2w FFE OutPhase coronal
FOV 150 mm
Voxel size 0.75 x 0.86 mm,
Slice thickness 3.5 mm
SENSE factor 1.5
Scan time 2:47 min.
FOV 150 mm
Voxel size 0.75 x 0.96 mm
Slice thickness 3.5 mm
SENSE factor 1.5
Scan time 2:17 min.
FOV 150 mm
Voxel size 0.65 x 0.77 mm
Slice thickness 1.5 mm
SENSE factor 1.6
Scan time 2:07 min.
T2w TSE transverse FOV 150 mm<br>
Voxel size 0.77 x 0.93 mm<br>
Slice thickness 3.5 mm<br>
SENSE factor 1.4<br>
Scan time 2:36 min.T2w SPAIR coronal FOV 150 mm<br>
Voxel size 0.85 x 0.98 mm <br>
Slice thickness 3.5 mm<br>
SENSE factor 1.5<br>
Scan time 3:02 min.T2*w FFE InPhase sagittal FOV 150 mm<br>
Voxel size 0.78 x 0.84mm<br>
Slice thickness 3.5 mm<br> SENSE factor 1.6<br>
Scan time 3:17min.
T2w TSE transverse
T2w SPAIR coronal
T2*w FFE InPhase sagittal
FOV 150 mm
Voxel size 0.77 x 0.93 mm
Slice thickness 3.5 mm
SENSE factor 1.4
Scan time 2:36 min.
FOV 150 mm
Voxel size 0.85 x 0.98 mm
Slice thickness 3.5 mm
SENSE factor 1.5
Scan time 3:02 min.
FOV 150 mm
Voxel size 0.78 x 0.84mm
Slice thickness 3.5 mm
SENSE factor 1.6
Scan time 3:17min.
STIR LongTE coronal FOV 150 mm<br>
Voxel size 0.75 x 0.84mm<br>
Slice thickness 3.5 mm<br> SENSE factor 1.7<br>
Scan time 3:02 min.WATSc coronal FOV 150 mm<br>
Voxel size 0.75 x 0.75mm<br>
Slice thickness 1.5 mm<br> SENSE factor 2.0 <br>
Scan time 2:54 min.WATSf transverse FOV 150 mm<br>
Voxel size 0.75 x 0.75 mm<br>
Slice thickness 1.5 mm<br>
SENSE factor 1.4<br>
Scan time 2:37 min.
STIR LongTE coronal
WATSc coronal
WATSf transverse
FOV 150 mm
Voxel size 0.75 x 0.84mm
Slice thickness 3.5 mm
SENSE factor 1.7
Scan time 3:02 min.
FOV 150 mm
Voxel size 0.75 x 0.75mm
Slice thickness 1.5 mm
SENSE factor 2.0
Scan time 2:54 min.
FOV 150 mm
Voxel size 0.75 x 0.75 mm
Slice thickness 1.5 mm
SENSE factor 1.4
Scan time 2:37 min.

Patient positioning

Position the patient head first and the coil loops in a perpendicular fashion so that the coil elements are as close to 90 degrees as possible. This enables SENSE to be used in all directions.

 

Use sand bags and other accessories to secure the coils and for patient comfort (not yet applied in the picture below).

 

Make sure the coil cables do not create any loops and that there is no skin contact with any part of the coil.

Optimal coil position One coil loop around the upper arm, the other parallel to the clavicle.Optional coil position One coil loop parallel to the upper arm, the other parallel to  the clavicle area.
Optimal coil position
Optional coil position
One coil loop around the upper arm, the other parallel to the clavicle.
One coil loop parallel to the upper arm, the other parallel to the clavicle area.


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
1644
Added:
Oct 9, 2008

Rate this:
Log in to vote
 

ExamCard
Panorama HFO
Release 2
Pulsar
Musculoskeletal, SENSE, Shoulder
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor