Panorama 0.6T CIA Shoulder for large patient

ExamCard
Wirtanen, Pirjo Philips Healthcare Philips Global

Download
Login is required to download ExamCards.

ExamCard purpose

This ExamCard is for shoulder examinations of large patients, whose shoulders do not fit in the standard shoulder coil. It uses CIA (Combined Image Aquisition) with Shoulder coil and Flex L+ coils.  R6.2 software and CIA license are needed.

 

 

ExamCard description:

This ExamCard contains protocols typically used in shoulder imaging. You can select the protocols required for your patient. Imaging times are about 5 minutes and SENSE is used in most scans.

 

All protocols can be use in any plane. However, remember to use horizontal phase encoding direction in coronal (L-R) and sagittal (A-P) scans, and vertical (A-P) in the transverse plane.

All protocols have over 200% phase foldover suppression, so you can just swap the phase.

 

This ExamCard has one scoutview, but you should rescan that as many times as needed to determine the right slice positioning. The position of the upper arm of the large patient is often compromised.

 

When imaging large patients with CIA coils, aliasing must be taken care of.

ExamCard overview

Knee for large patient
Scan 13 Plane Scoutview
Scan 2T2w TSE transverse
Scan 3PDw TSE transverse
Scan 4T1w TSE coronal
Scan 5T2w IR-TSE coronal
Scan 6PDw IR-TSE coronal
Scan 7T2w FS-TSE transversal
Scan 8T2w FS-TSE coronal
Scan 9PDw FS-DRIVE transversal
Scan 10T1w FS-SE coronal
Scan 11T2'w PR-FFE sagittal

ExamCard images

PDw TSE FOV 260 mm, <br>slice thickness 3.0 mm, <br>matrix 324, SENSE factor 1.6,<br>
scan time 4:54 min.PDw FS-DRIVE FOV 260 mm, <br>slice thickness 3.0 mm, <br>matrix 384, SENSE factor 1.5,<br>
scan time 4:46 min.
PDw TSE
PDw FS-DRIVE
FOV 260 mm,
slice thickness 3.0 mm,
matrix 324, SENSE factor 1.6,
scan time 4:54 min.
FOV 260 mm,
slice thickness 3.0 mm,
matrix 384, SENSE factor 1.5,
scan time 4:46 min.
T2w TSE FOV 260 mm, <br>slice thickness 3.0 mm, <br>matrix 260, SENSE factor 1.3,<br>
scan time 5:42 min.T1w TSE FOV 260 mm, <br>slice thickness 3.0 mm, <br>matrix 294, SENSE factor 1.5,<br>
scan time 5:35 min.T1w FS-SE FOV 260 mm, <br>slice thickness 4.0 mm, <br>matrix 260, <br>
scan time 5:33 min.
T2w TSE
T1w TSE
T1w FS-SE
FOV 260 mm,
slice thickness 3.0 mm,
matrix 260, SENSE factor 1.3,
scan time 5:42 min.
FOV 260 mm,
slice thickness 3.0 mm,
matrix 294, SENSE factor 1.5,
scan time 5:35 min.
FOV 260 mm,
slice thickness 4.0 mm,
matrix 260,
scan time 5:33 min.
T2w FS-TSE FOV 260 mm, <br>slice thickness 4.0 mm, <br>matrix 320, SENSE factor 1.6, <br>
scan time 5:36 min.T2w IR-TSE FOV 260 mm, <br>slice thickness 4.0 mm, <br>matrix 346, SENSE factor 1.4, <br>
scan time 5:04 min.T2*w PR-FFE FOV 260 mm, <br>slice thickness 4.2 mm, <br>matrix 346, SENSE factor 1.4, <br>
scan time 5:19 min.
T2w FS-TSE
T2w IR-TSE
T2*w PR-FFE
FOV 260 mm,
slice thickness 4.0 mm,
matrix 320, SENSE factor 1.6,
scan time 5:36 min.
FOV 260 mm,
slice thickness 4.0 mm,
matrix 346, SENSE factor 1.4,
scan time 5:04 min.
FOV 260 mm,
slice thickness 4.2 mm,
matrix 346, SENSE factor 1.4,
scan time 5:19 min.

Shimming procedure for fatsat

When using the CIA method with a large patient and a large FOV, special attention for fatsat imaging is needed. This ExamCard has a special procedure to optimize fat suppression:

 

  1. Coil change for FatSat Shims
    Default coil is the Shoulder coil, but you can change that by double clicking the item and selecting "properties", if Flex L+ coil gives better signal.
  2. Shimming with one coil
  3. Coil change for FatSat scans
    Both coils are activated.

Positioning

Position the patient "Head left". Use "Low" table position to ensure the patient comfort.

 

Coil positioning:  Shoulder coil as high as it goes

                         Flex L+ coil around the collum

                         Flex L+ coil inside the Shoulder coil, then SENSE can be used in all directions

Coil and patient positioning CIA coil positioning CIA coil positioning
Coil and patient positioning
CIA coil positioning
CIA coil positioning


This content has been made possible by NetForum Community.
Share this on: Share your link in twitter Share your link in facebook Share your link on LinkedIn Print Rate this article: Log in to vote

 
Rating:
Votes:
0
Views:
1158
Added:
Jun 30, 2006

Rate this:
Log in to vote
 

ExamCard
Panorama 0.6T
Release 6
Standard
Musculoskeletal, Shoulder
 

Clinical News
Best Practices
Case Studies
Publications and Abstracts
White Papers
Web seminars and Presentations
ExamCards
Protocols
Application Tips and FAQ
Training
Try an Application
Business News
Case Studies
White Papers
Web Seminars and Presentations
Utilization Services
Contributing Professionals
Contributing Institutions
Become a Contributor